Mon Mar 30 16:25:05 2020 |
ruijiu | Detectors | turn down the HV | turn down the hv and pull out the detector.
suggest by
Hi everybody,
is there now much more rate on the detectors?
Perhaps it helps to turn down the bias for a while and to take the detectors out of the beam, by pulling the pocket out?
And then see if the leakage current comes down again.
good luck, Thomas
***********************************************
Dr. Thomas Faestermann | Tel: +49-89-2723868
Physik Department E12 |
Techn. Univ. Muenchen | Mob: +49-1626193388 |
103
|
Mon Mar 30 20:51:06 2020 |
Ragan | Detectors | Bias Voltage for Si-pads | The limit of the leakage current for all Si pads was increased from 1500 nA to 5000nA and
leakage current was recorded for different Bias Voltages (32V, 35V, 40V, 45V, 50V, 55V
and 60V).
Since, I_leakage remains < 3 uA for 60 V, it is decided to increase the Bias Voltage for
all the Si-pads from 32 V to 60 V.
For DSSD and CsI, bias voltages remains same (-50V and -160V respectively). |
111
|
Wed Apr 1 01:02:24 2020 |
ruijiu | Detectors | the detector cable is disconnected. | |
112
|
Wed Apr 1 01:05:36 2020 |
ruijiu | Detectors | detector position fixed at 60 mm | The detector cable was disconnected . We entered and reconnected the cable. |
178
|
Thu Nov 19 09:14:40 2020 |
Jan | Detectors | remote MSCF settings for CsISiPHOS | Attached is a list of all set values for the 5 MSCF shaper modules.
This was done prior to run0251, which served as a final benchmark for those settings.
See: https://elog.gsi.de/esr/E121/62
And https://elog.gsi.de/esr/E121/104
The det. to MSCF assignment is the following:
MSCF 1: Si-pad 1 + 2 p-side/strips (ch. 1-7 + ch. 9-15)
MSCF 2: Si-pad 3 + 4 p-side/strips (ch. 1-7 + ch. 9-15)
MSCF 3: Si-pad 5 p-side/strips (ch. 1-7)
MSCF 4: Si-pad 6 p-side/strips (ch. 1-7)
MSCF 5: DSSD (ch. 1 - 4); CsI (ch. 7 & 8); Si-pad n-sides (ch. 11 - 16)
Note that only the pre-amplification stage of the MSI-8 amplifiers were used for all channels. |
180
|
Tue Dec 8 15:47:07 2020 |
Jan | Detectors | manual preamp & shaper settings for CsISiPhos | attached is a document showing fotos of all involved amplifiers in the chain to document the manual settings active during the experiment.
Summary:
Si-pad n-sides [MSI-8 (1)] were set to 5 GeV (expected E-deposition ~2 GeV)
Si-pad p-sides [MPR-32 (1 & 2)] were set to 1 GeV (expected E-deposition ~2 GeV)
DSSD channels [MSI-8 (2)] were set to 5 GeV (expected E-deposition ~1.2 GeV)
CsI channels [MSI-8 (2)] were set to 4 GeV (expected E-deposition 46 GeV)
It seems we got clipped signals from the Si-pad p-sides, because the MPR-32 were saturated due to a wrong gain setting.
This might explain, why there was not real influence on the spectra when adjusting MSCF settings for those channels.
I cannot comment on the CsI gain setting, because i don't know whether the given GeV-range by MesyTec directly applies for CsI & photodiode.
However, for DSSD and Si-pad n-sides the settings look okay. |
2
|
Mon Dec 3 15:06:04 2018 |
Chen, Sidhu | DAQ | Test of two preamplifier. | Description:
We have two pre-amplifiers:
New: Max. range 5GeV, Two sensitivity options: 1 GeV and 5 GeV, Decay time = 30 us, pulse input 220 MeV/u, Ser. #MPR-32-D-5000-30-1118380.
Old: Max. range 5GeV, Two sensitivity options: 1 GeV and 5 GeV, Decay time = 25 us, pulse input 220 MeV/u. Ser. #MPR-32-D-5000-25-0414310.
Conclusion:
Both pre-amplifiers work well.
The summary is attached in the followings:



 |
7
|
Tue Mar 12 12:54:58 2019 |
Chen, Sidhu | DAQ | how to start DAQ and plot histogram. | 1. login litv-exp account
rchen@lxg1254:/data.local2/2002_238UIMS$ ssh -X litv-exp@lxg0155.gsi.de
(don't login on lx-pool.)
litv-exp@lx-pool.gsi.de's password: ViValasVegas@****
2.login r4l-47 account
litv-exp@lxi087:~$ ssh r4l-47
litv-exp@r4l-47's password:ViValasVegas@****
3.start mbs
R4L-47 litv-exp > mbs
mbs>
mbs> -R4L-47 :msg_log :Message logger running
3.1 Initilizing, show, start and stop the acq
mbs> @startup
-R4L-47 ispatch :Task "m_util" not started!
mbs> @startup
-R4L-47 :util :task m_util started
-R4L-47 :util :setup file setup.usf successfully loaded
-R4L-47 :util :trigger module set up as MASTER, crate nr: 0
-R4L-47 :util :enabled interrupt
-R4L-47 :collector ipe type 2 or 4: virtual mapping
-R4L-47 :transport :task m_transport started
-R4L-47 :read_meb ipe type 2 or 4: virtual mapping
-R4L-47 :transport :starting server in inclusive mode
-R4L-47 :transport :waiting for client (port 6000)
mbs> -R4L-47 :read_meb :task m_read_meb started
-R4L-47 :stream_serv:task m_stream_serv started
-R4L-47 :collector :task m_collector started
mbs> show acq
-R4L-47 ispatch :Task "m_util" not started!
mbs> show acq
-R4L-47 :util :--------------------------------------------------------------
-R4L-47 :util :ACQUISITION INITIALIZED 12-Mar-19 13:44:09: active tasks:
-R4L-47 :util :m_util m_transport m_daq_rate m_collector m_stream_serv
-R4L-47 :util :m_read_meb
-R4L-47 :util :SB setup LOADED from setup.usf.
-R4L-47 :util :Readout table NOT LOADED, acquisition STOPPED.
-R4L-47 :util :There are 32 streams with 32 buffers a 32768 each.
-R4L-47 :util :Current crate is 0, event builder is RIO4.
-R4L-47 :util :--------------------------------------------------------------
-R4L-47 :util :Crate 0: ID= 20, LOCAL SYNC SYNC , subevent slots=200000.
-R4L-47 :util : trig 1: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 2: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 3: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 4: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 5: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 6: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 7: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 8: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 9: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 10: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 11: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 12: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 13: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 14: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util : trig 15: cvt= 300, fclrt= 12 [usec], max se length=100000 b
-R4L-47 :util :--------------------------------------------------------------
-R4L-47 :util :Name of output device =
-R4L-47 :util :Tape label = , Dismounted, file CLOSED
-R4L-47 :util : 0.000 [MB] written to tape, 0.000 to file
-R4L-47 :util :--------------------------------------------------------------
-R4L-47 :util :Collected: 0.0000 MB, 0 Buffers, 0 Events.
-R4L-47 :util :Rate : 0 KB/s, 0 Buffers/s, 0 Events/s
-R4L-47 :util :--------------------------------------------------------------
mbs> start acq
mbs> stop acq
-R4L-47 :util :stop acquisition
mbs> -R4L-47 :collector :acquisition NOT running
-R4L-47 :read_meb :found trig type 15 == stop acquisition
4. plot the histogram
litv-exp@lxi087:~/2019Munich-Det/Go4-Muenchen-3V785AH-1V830-simple$ go4
Go4 v5.3.2, build with ROOT 6.12/06 and Qt 4.8.6
Using Qt settings at /u/litv-exp/.config/GSI/go4.conf (default)


 |
9
|
Tue May 28 08:36:14 2019 |
Jan | DAQ | MBS start & RFIO server | You need a DAQ-PC and the RIO crate computer.
All of the DAQ proceedures laid out here are operated from inside the DAQ screen session.
Right now for e121 the following DAQ-setup is used: (might change in the future)
DAQ-PC: lxg0155
screen session: e121_daq @ lxg0155
RIO: r4l-47
DAQ dir: /esr/usr/litv-exp/2019Munich-Det
The most important commands:
on DAQ-PC
mbs - start mbs DAQ
rawDispRFIO64 - start RFIO server
in MBS shell
@startup - initialize MBS (only inside DAQ dir)
@connect - connect RFIO server (only if server running)
@runstart - open file for data writing
@runstop - close file
start acq - start DAQ
show acq - status of DAQ
show file - status of current file/RFIO server
The full proceedure for the DAQ:
- on DAQ-PC enter screen session:
litv-exp@lxg____:~$ screen -x e121_daq or create new screen session if non is exisiting:litv-exp@lxg____:~$ screen -S e121_daq - inside screen: go to 1st tab (MBS) and login to RIO if not yet done:
<Ctrl-a 0>
litv-exp@lxg____:~$ ssh r4l-47
- on RIO reset and start MBS
R4L-47 litv-exp > cd 2019Munich-Det
R4L-47 litv-exp > resl
R4L-47 litv-exp > mbs
mbs > @startup
- inside screen go to 2nd tab (RFIO)
<Ctrl-a 1> or create 2nd tab if not exisiting<Ctrl-a c> - start RFIO server on DAQ-PC
litv-exp@lxg____:~$ cd /u/litv-exp/rfio
litv-exp@lxg____:~$ ./rawDispRFIO64-wheezy
-I- 64 bit RFIO test server (lxg0155) listening on port 1974
forks './rawServRFIO64-wheezy' for each client request - inside screen go to 1st tab and connect MBS to RFIO server
<Ctrl-a 0>
mbs > @connect
mbs> show file
mbs> -R4L-47 :transport :Name of output device = RFIO
-R4L-47 :transport :current tape label = lxg0155 , opened
-R4L-47 :transport :current file name = , closed
-R4L-47 :transport : 0.000 MBytes written to tape
-R4L-47 :transport : 0.000 MBytes written to current file - start acquisition
mbs > start acq
mbs > show acq
...
-R4L-47 :util :Name of output device = RFIO
-R4L-47 :util :RFIO server lxg0155, file CLOSED
-R4L-47 :util : 0.000 [MB] written to tape, 0.000 to file
-R4L-47 :util :--------------------------------------------------------------
-R4L-47 :util :Collected: 61.1451 MB, 1866 Buffers, 131990 Events.
-R4L-47 :util :Rate : 33 KB/s, 1 Buffers/s, 71 Events/s
-R4L-47 :util :-------------------------------------------------------------- - open file / start a run
mbs > @runstart
-R4L-47 :transport :open file at server lxg0155::
-R4L-47 :transport :/data.local1/E121/predata/e121_run_1239.lmd - close file / stop a run
mbs > @runstop
-R4L-47 :transport :Rfio server lxg0155: closed file after 0.262 MB written:
-R4L-47 :transport :-I- remote output file /data.local1/E121/predata/e121_run_1239.lmd closed
mbs> -R4L-47 :transport :/data.local1/E121/predata/e121_run_1239.lmd
|
10
|
Wed Jun 26 16:20:31 2019 |
Jan | DAQ | unpacker: lmd > root | I have written an unpacker for the lmd data taken with the current daq-system for E121.
The crate configuration is as follows:
ADC1 channels 1-32 (Caen 785)
ADC2 channels 1-32 (Caen 785)
ADC3 channels 1-32 (Caen 785)
SCALER channels 1-32 (Caen 830)
The unpacker binary can be found here
/u/litv-exp/unpacker/unpackexps/E121/E121
The current spec file is attached.
To convert lmd-files to root-files, please use it as follows:litv-exp@lxg___: E121_unpacker /path/to/file.lmd --ntuple=RAW,/path/to/new_file.root |
15
|
Tue Feb 25 15:48:31 2020 |
ruijiu chen | DAQ | Important folders of e121 eperiments: | 1. slow control:
1.1. program file:
litv-exp@ringpi:~/epics/mrcc_ioc/db $
1.2 user interface:
litv-exp@lxg1298:~/E121/scripts/GUIe121_new
2. mbs:
2.1 user interface
litv-exp@lxg1298:~/E121/scripts$ ./e121_gui.py
2.2 source
R4L-47 litv-exp > pwd
/esr/usr/litv-exp/2020_e121/
3. go4:
litv-exp@lxg1298:~/2019Munich-Det/Go4-Muenchen-3V785AH-1V830-simple$ go4
4. experimenal data:
litv-exp@lxg1298:/data.local1/E121 |
16
|
Tue Feb 25 15:49:58 2020 |
ruijiu chen | DAQ | changing scaling factor. | How to change scaling factor?
The parameters is in f_user.c file:
1. open the f_user.C with emacs :
R4L-47 litv-exp > emacs -nw f_user.c
2. change the parameters "MWPC_downscale".
#define MWPC_downscale 0 // reduction = 2**MWPC_downscale
3.compile the program.
4. restart mbs. |
33
|
Sat Mar 28 12:19:21 2020 |
Ragan, RuiJiu | DAQ | Change of the polarity of Amplifiers | Since last night, the signals from the p-side of all the Si pads were seen as negative on picoscope.
The signals from DSSSD, all n-sides of Si pads and CsI were positive which all are in MSCF-5.
Thus, I and RuiJiu, changed the polarity jumpers of MSCF 1,2,3 and 4(all the four jumpers inside the mentioned MSCF) and now hope to see positive signals from the p-side of all the Si pads. |
37
|
Sat Mar 28 15:07:40 2020 |
ruijiu | DAQ | start run | |
41
|
Sat Mar 28 16:12:25 2020 |
ruijiu, ragan | DAQ | enter ESR and turn on the HV crate. | |
44
|
Sun Mar 29 09:40:35 2020 |
ruijiu | DAQ | restart go4 | go4 crush up.
the problem is
due to mbs.
restart mbs.
see an error.
restart again.
the mbs and go4
are working
working. |
51
|
Sun Mar 29 13:05:36 2020 |
ruijiu | DAQ | start ntcap | IQ_2020-03-29_13-05-
56
SC_2020-03-29_13-05-
56 |
98
|
Mon Mar 30 14:24:46 2020 |
ruijiu | DAQ | stop taking data. | 1. stop run0272.lmd.
2. stop NTCAP.
NTCAp file:
SC_2020-03-30_09-28-14
IQ_2020-03-30_09-28-14
(saved in drive: p) |
175
|
Thu Oct 1 22:20:03 2020 |
Jan | DAQ | final unpacker config | We identified an issue with randomly missing adc channels in all lmd-files before run 0273.
This led to errors in the unpacking process.
This issue is solved with the new config and mapping for the unpacker.
The respective files are attached here.
The unpacker is ready to use in /u/litv-exp/unpacker/unpackexps/E121
The adc issue is also understood. We deactivated the internal adc thresholds during the experiment, because we encountered the very same unpacker errors. This solved the problem for all later runs, but we forgot to think about a solution for the earlier runs.
Mea culpa.
Jan |
176
|
Mon Nov 2 16:45:55 2020 |
Jan | DAQ | f_user & mbslog | Attached is the f_user.c and the mbslog.l files.
The readout function is for the following setup:
3x ADC CAEN V785
1x SCA CAEN V830
1x TDC CAEN V1190
1x Vulom TRB5 (triggerbox)
1x Vulom4b PEV (priority encoder)
1x TRIVA7
1x RIO4 R4L-47 |
|